I’m having trouble figuring which step I am going wrong in jMRUI when trying to quantify my choline peak. Below is some background and what I have been attempting:
I have acquired spectroscopy data from a choline phantom (water + choline) using a Bruker system and I am using this data to attempt to quantify choline. When I create my metabolite list (using NMR Scope-B) and I load the metabolite list in 1D mode, the metabolites are in the correct location (ppm). However, when I attempt to fit a metabolite to acquired data, the metabolites shift and are no longer in their proper location. This happens even though I calibrate the frequency axis of the acquired spectrum. I should mention that in NMR Scope-B, I am trying to manually input the pulse sequence which I don’t have much knowledge about so perhaps that’s where the fault is. How important is the pulse sequence for the metabolite list? Could this be why my metabolite peaks are shifted once I attempt to overlay it onto my spectrum in quest?
Also, I seem to be missing where the example data sets are?
Thank you,
Gena
Note: I’ve tried both jMRUIv5.2 and jMRUIv6.0